I have had this tube tester for like 25 years now. Just did a calibration based on Bob Putnak's webpage about the 648 series. Did the solid state conversion first, then ran through the various test procedures. All went well! I still have 2 issues with the tester. First, the Life Line switch is malfunctioning. For those that are not familiar with it, it is a momentary toggle (SPDT) which drops the filament voltage 10% when the switch is thrown. It had worked in the past, but now it just goes open when I hit the toggle - heater voltage drops to zero. If anyone can point me to a proper replacement, please do. The other issue was the leakage test. Bob said use the shorts test setup, but instead use a 1 Meg resistor in place of a short. Should see the shorts light dimly on at the 1 Meg setting, and get brighter as you advance to 2 Meg. But I don't see the glow at 1 Meg, and it's barely there at 2 Meg. There is a 220K resistor in series with that shorts potentiomer and I was wondering if dropping it to 180K or even 150K would alleviate this issue? Bob did say to look at the pot and resistor if there were any out of the ordinary issues, but the pot does measure 1 Meg per print, and the 220k measures 223K - not far off. How to proceed? All else is well with this tester - plate, screen and grid voltages all line up as per testing.
-Bob-
-Bob-