Chassis leakage current measurement setup?

whoaru99

Epic Member
Recent thread about getting a tingle when touching chassis got me thinking a bit more about this. Have seen versions of this in various service manuals for checking for leakage current

Here's a specific example from Yamaha -

1732899110989.png

Does the following drawing correctly indicate the test setup described in the Yamaha process, "meter impedance equivalent to 1500 ohms shunted by 0.15uf"?

1732899348578.png
 
Register to hide this ad
The figure below is from an electrical safety testing white paper from Fluke. Leakage current measuring devices use this impedance circuit for measurements.

Meter impedance.jpg
 
I appreciate there may be more than one way to do it.

But, the question is specifically about whether or not the drawing reflects the test setup described by Yamaha.
 
Yamaha says to measure current with a meter having an impedance of 1.5k Ohms in parallel to 0.15uF.

The "1.5k || 0.15u" represents an analog of a human body.

Ideally the measurement should include the current through the capacitor. However, (1) often interest is only on 60 Hz and not very much 60 Hz goes through the cap, and (2) AC current meters often don't properly measure current (or necessarily have zero impedance) above a few hundred Hz so any HF is not measured properly anyway.
 
Yamaha says to measure current with a meter having an impedance of 1.5k Ohms in parallel to 0.15uF.

Which is what the drawing represent, correct?

Since a multimeter's current function typically has very low inherent resistance the series resistor is what makes it 1500 ohms net, then the parallel/shunt capacitance.
 
Curious, would Fluke's interpretation be sufficient, or is it lacking?
The tests given by the audio equipment mfgs in these old manuals seem to be derived from a particular UL standard that is no longer current. The Fluke circuit probably derives from some different set of standards entirely (but obviously with similar purpose).
 
The paper linked in an earlier post shows a circuit like that in the Fluke drawing. I believe it mentions the standard(s) where applicable (and a whole bunch more).
 
1500 Ohms and 0.15 uF in parallel with an RMS voltmeter (though an average-reading one won't be noticeably different). Current is voltage measured, divided by 1500 - 1.5V is 1.0 mA. This setup is mentioned in numerous safety standards, though some current ones show the network shown by Fluke (should give the same results with less math (1V is 1 mA) Yes, this is to represent a typical human body; capacitor rolls off higher frequencies, since they are less hazardous than 60 Hz.
 