So I try to do Gm and I went through this and I am unable to get ANY movement out of the meter while it is in shorts test. If I switch it over to GM-emissions it instantly flies to the 100% where I can further adjust it using R24. Is this correct?Jay,
This is from my notes.
Here is my 10 steps GM calibration procedure.
CALIBRATION (After the Emission Calibration 60 is properly done).
1) Set a sensitive voltmeter to AC.
2) Set the FUNCTION switch to SHORT, this power on the tester and warm up the 12AU7.
3) Add the 1.6k resistance between the wiper of SIGNAL F pot and the Cal point on the board.
4) Set the B LOAD pot to position A.
5) Set the SIGNAL F to maximum 100.
6) Adjust R24 to reach 100%. Sometimes you need to adjust R22 for fine tuning so you're
not at the edge of R24 (if made adjustable during the PCB build).
7) Measure the AC voltage at the SIGNAL wiper and ground on the PCB: I read 0.722VAC.
8) Set the SIGNAL F to minimum (0) position.
9) Disconnect the 1.6K resistor.
10) Measure the AC voltage at the SIGNAL wiper and ground on the PCB and adjust R21
1/100 of the previous voltage, in my case 7.2mV.
That's it. If you have a scope you can do it like this too: you just have to have 1/00 of the signal amplitude at the end. Same principle but just visual.
Hello,Hi Guys,
I am not sure this belongs to the Tube Audio forum but I didn't find a more suitable category.
I worked of this tester, a Sencore MU-150, that I found very nice so I am sharing this with you.
The good points:
- This tester is one of the last tube tester been produced, therefore is more modern than the classic oldies.
- It has a very sensitive as it uses a 12AU7 tube as voltage differential. The meter could pick up variation that my fluke couldn't.
- It does the classic tests for static emission, shorts, grid leakage and tube life.
- And does a superb job as trans-conductance test, by generating a 5 Khz signal to the tube and measure it that way. It has 3 transistors for that function.
- The voltage reference is also made with a tube too, an OB2 voltage regulator.
- It is a very robust and heavy piece of equipment. It has 2 transformers, one for the heaters and one for the tester itself. The case is a made of solid metal protecting the tester very well. The layout is clean and well design.
The weak points:
- The main problem is the PCB. Very brittle and prone to cracks. It attached directly to back the meter.
- The calibration is unnecessary complicated.
Here are some pictures.
View attachment 1477235
Inside:
View attachment 1477236
Here is the schematic:
View attachment 1477239
Most of the components on the board were out of specs and had to be changed.
All external caps were bad.
But all the rest was in excellent condition.
So I first tried to change all components on the board and give it a try.
Here it is:
View attachment 1477241
The tester was working. I used a SS voltage regulator too.
I was happy, but the board was still fragile, and the calibration wasn't easy as some resistor needed adjustment and without a trimmer, it the painful de-soldering/re-soldering process, try, retry.
I liked this tester so I decided to make a new board, and improve a few things.
- First I wanted all the caps to be mounted on the board and not externally mounted to the chassis.
- Increase within reason the PS filtration.
- I wanted to only use 1 tube, and replace the OB2 with a solid state solution.
- I wanted to shield the 12AU7 and keep it tied to the socket.
- Add trimmers were I needed to make the calibration easy.
- I added a current sensing resistor in the meter path so I could connect a VDM to it.
After 2 iterations, one without layout change (V1.0) and all these changes (V1.1), I got a perfectly working solution.
Here it is:
View attachment 1477251
Only one tube mounted on a plate, shielded. All caps in the board, more trimmer, neon bulb on top, etc..
Here is it finished.
View attachment 1477279
The calibration was easier.
1) The static emission needs a plug to be build with 2 resistors and a diode, as per the manual. Then you insert it and adjust the meter mid range with 1 trimmer. That calibrates the tube voltage differential for all measurement.
2) The GM emission calibration is a 2 step process. You need a 1600 ohm resistor to the arm of the signal pot, to the 0.47uf cap of the voltage amplifier (marked CAL in to board), as the user manual's instruction.
- Max signal first, you adjust the meter at 100% with 2 trimmers. Then forget about the scope things in the manual. Get a DVM and measure the voltage at the arm of the signal pot
- Min signal, remove the 1.6 resistor and adjust the voltage to 1/10 of what you measured before. That part is tricky.
And that's it. In fact they wanted the max signal to be 100% of the meter, and the min a tenth of it to have a linear range.
I added a fuse for safety, changed the neon bulbs (nice one from Radioshack), and all load resistors mounted on the rotary switch. Nothing difficult.
I am sure I am forgetting things but that's mostly it.
I like the ease of using of that tester. The little book has most of the tubes in it, you start by the shorts, then emission, then grid leakage, then gm emission then life test. Simple.
I attached my gerber file so anyone can have a PCB made for it.
And here is the rendered picture.
Thanks for reading.
View attachment 1477278

I have another question. The board I have I bought off of Ebay. V1.1. But I was going by your download files above which included an xls sheet of mouser parts. The spreadsheet seems to be different in some cases than the part values on the board. For example on the spreadsheet For C1, C2A, C2A-2 the board says 82uf 250v. But the spreadsheet says 100uf 250 volts. I assume that the spreadsheet parts list will definitively work for this project? There are also some item quantity issues for example it says 2 rectifiers are needed when only 1 is needed. And shows 10 36v 5W and 10 35v 3W. I just want to make sure I have everything correct before placing my order. Thank You! -PeteHi Pete,
Unfortunately, there is no off the shelve solution for this as I couldn't fine 90 degree socket anymore.
So I made a simple plate to mount the socket to it, and I mounted the plate with 90 degree brackets to the board.
There are 2 holes on the PCB for the brackets.
It's not fun to do, as I don't know for you, but for me the mechanical part is always where I mess up.
But at the end, it worked well, and I had zero issue with my tester for 5 years now already. And I use is quite often.
Also some other AK'er used it so it had some use, transportation and such.
Only a 8 pin Chinese socket saver broke, ha ha.
I hope it helps, and good luck with your board.
Brice.
Here is the picture for reference.
View attachment 3307399
Got it! I am super paranoid so just wanted to ask. Thank you so much for the response!Hi Pete,
V1.1 seems like my board, yes.
C1,C2A and C2A-2 are power supply caps, so here the bigger the better.
What's important is the caps lead spacing and diameter to fit on the board, here 7.5mm LS and 16mm D, but also the length of 20mm max else it may touch the enclosure.
100uF is good.
The spreadsheet is old and most likely not accurate anymore. I don't know if all the parts are still listed.
Anyway, the 10x is most likely my purchase order as I often buy parts by 10 as so cheap.
Check the board for exact quantity as at the end of the day, you need to populate that board.
Sorry I just went through every post and could have answered most of my questions. I am now putting together the parts list to order. I was looking at CR12 and see there are two in series 1N5817 and 1N4007. Does position matter for those? On the ebay listing for the board they show CR12B (1N5817) and CR12A (1N4007) but the board has both as CR12. Thank You!Hi Pete,
V1.1 seems like my board, yes.
C1,C2A and C2A-2 are power supply caps, so here the bigger the better.
What's important is the caps lead spacing and diameter to fit on the board, here 7.5mm LS and 16mm D, but also the length of 20mm max else it may touch the enclosure.
100uF is good.
The spreadsheet is old and most likely not accurate anymore. I don't know if all the parts are still listed.
Anyway, the 10x is most likely my purchase order as I often buy parts by 10 as so cheap.
Check the board for exact quantity as at the end of the day, you need to populate that board.