Hi Dog,
I've just re-calibrated mine and took notes as I get calibration questions quite often.
First make sure you have a good voltmeter on hand. Many DVM can't really integrate accurately this square signal.
My Fluke 117 can't, it just displays 0. I used my bench Tektronix for that.
And I set my Tester that way to have easy access to both sides:
And here is a picture of my board so you can see the pots I have:
Here is my 10 steps GM calibration procedure.
CALIBRATION (After the Emission Calibration 60 is properly done).
1) Set a sensitive voltmeter to AC.
2) Set the FUNCTION switch to SHORT, this power on the tester and warm up the 12AU7.
3) Add the 1.6k resistance between the wiper of SIGNAL F pot and the Cal point on the board.
Calibration point on the board:
4) Set the B LOAD pot to position A.
5) Set the SIGNAL F to maximum 100.
6) Adjust R24 to reach 100%. Sometimes you need to adjust R22 for fine tuning so you're
not at the edge of R24 (if made adjustable during the PCB build).
7) Measure the AC voltage at the SIGNAL wiper and ground on the PCB: I read 0.722VAC.
8) Set the SIGNAL F to minimum (0) position.
9) Disconnect the 1.6K resistor.
10) Measure the AC voltage at the SIGNAL wiper and ground on the PCB and adjust R21
1/100 of the previous voltage, in my case 7.2mV.
That's it. If you have a scope you can do it like this too: you just have to have 1/00 of the signal amplitude at the end. Same principle but just visual.
In your case, I think you have a mechanical or wiring problem, and I hope you find it soon.
I hope it helps.
Brice.